Hi Everyone,

This morning the Oxygen depletion alarm in -1.34B (TIMS room) was alarming, showing 28.3% O2. Given we have no Oxygen cylinders in the lab, this is clearly a faulty sensor.  I've inhibited the sensor from the alarm system to enable us to quieten the alarm, and the engineer will attend to swap out the faulty sensor tomorrow morning.

The lab is open for use, but please treat any other alarms as real alarms, until I can confirm otherwise.

Cheers,
Lindsey